RE: cfm to slpm
Hi Jim
Understanding gas flow and measurement in thermal spray processes does seem to cause a lot of confusion, particularly when trying to relate gas flows between different control systems for operation of the same gun, like trying to duplicate parameters between 7MC, 9MC and A3000 etc. for say a 9MB gun.
First point, we are considering a volumetric measurement per amount of time of gas. Problem here is that, what is important to the thermal spray process, is the AMOUNT of gas (number of atoms/molecules or mass). A volumetric measurement of a gas is really meaningless by itself, unless we consider the effects of temperature and pressure (same number of atoms/molecules or mass of a gas will ocupy different volumes if we change conditions of temperature or pressure). So it is important that if volumetric flow is used to measure/contol the amount of gas, then conditions need to be known and flow units standardised. Common units used:
* Metco Flow Meter Reading (FMR) (specific to equipment/process)
* SCFH (Standard Cubic Feet per Hour (70F)
* N.L.P.M or SLPM (Normal/Standard Litres per Minute (0C)
If we consider say an older 7MC system, which uses flow meter tubes and reads in Metco Flow Meter Reading (FMR) units. These units are practically meaningless outside of the equipment/material spray parameters designed for it. Conversion to standard units is possible knowing FMR, pressure, temperature and tube calibration conditions (see here) Hydrogen secondary gas is far more complicated I won't explain here (see here). More recent equipment either use flow meters (using specific parameters) designed to give direct readout in SCFH or SLPM/NLPM or mass flow meters.
Quote:Looking at conversion charts 1 cfm = 28.3 lpm thus a typical gas flow of 100 cfm = 2830 lpm right? This doesn't seem right since the A3000 only allows 100 slpm! Am I missing something with the standard part of standard litres per minute?
Yes, 1 cfm = 28.3 lpm thus a typical gas flow of 100 cfm = 2830 lpm
Are you sure you not confusing CFM with CFH? 100 SCFH is in the right ball park for plasma gas flow. 1 SCFH = 0.472 LPM so 100 SCFH = 47.2 LPM - more like it?
but wait, 47.2 LPM from above calculation really means 47.2 LPM @ 70F around normal atmosheric pressure (standard conditions for SCFH). This is not the same as SLPM/NLPM where standard conditions are 0C not 70F . So we need to apply corrections for the different conditions. So 100 SCFH = 47.2 LPM = 43.8 SLPM or NLPM.
You can use the plasma gas flow calculator to do this easily by inserting say 100 value into one of the SCFH boxes, calculate and read across.
Final point, you need to be careful where your gas flow parameters are coming from and you need to know what equipment and gun (and set-up) they are intended for. For example you can not use 3MB/7MB/9MB gas flow parameters say for a F4 gun or for that matter when using different nozzles.
Sorry for long lecture just trying to cover everybody. Hope it helps.
|
|
06-07-2008, 12:15 AM,
|
|
hvofhamid
Senior Member
|
Posts: 95
Threads: 9
Joined: Dec 2006
Reputation:
2
| |
RE: cfm to slpm
TurbineRepair Wrote:Hi all,
Could anyone advise me how to convert cfm to slpm?
I'm running a A3000 Metco system on with a S3 robot. I'm spraying PAC and Praxiar powders and I believe they quote their gas flow in cfm.
Looking at conversion charts 1 cfm = 28.3 lpm thus a typical gas flow of 100 cfm = 2830 lpm right? This doesn't seem right since the A3000 only allows 100 slpm! Am I missing something with the standard part of standard litres per minute?
For the flow above (Praxiar Inco 718 powder Argon flow - 3M) the guys have the machine set to 55 slpm. I've no clue how they came to this figure.... taking to them they don't either!
Thanks
Jim hello
please can you explain the terms :
cfm?
lpm,?
and splm?
i tried to understand but terms can means several things,of course I UNDERSTAND YOU but young people must undestand what you say,in this case you have to give them complete meaning of abreviations
may be they have solution for you and me
|
|
06-07-2008, 01:42 AM,
|
|
Gordon
Administrator
|
Posts: 1,802
Threads: 70
Joined: Jun 2006
Reputation:
9
| |
RE: cfm to slpm
Hi Hvofhamid
Sorry if the meaning of the abreviations were not clear
CFM - Cubic Feet per Minute
LPM - Litres per Minute
FMR - Metco Flow Meter Reading
SCFH - Standard Cubic Feet per Hour
N.L.P.M or SLPM - Normal/Standard Litres per Minute
|
|
10-15-2008, 08:33 AM,
|
|
RE: cfm to slpm
(06-06-2008, 12:56 PM)Gordon Wrote: Hi Jim
Understanding gas flow and measurement in thermal spray processes does seem to cause a lot of confusion, particularly when trying to relate gas flows between different control systems for operation of the same gun, like trying to duplicate parameters between 7MC, 9MC and A3000 etc. for say a 9MB gun.
First point, we are considering a volumetric measurement per amount of time of gas. Problem here is that, what is important to the thermal spray process, is the AMOUNT of gas (number of atoms/molecules or mass). A volumetric measurement of a gas is really meaningless by itself, unless we consider the effects of temperature and pressure (same number of atoms/molecules or mass of a gas will ocupy different volumes if we change conditions of temperature or pressure). So it is important that if volumetric flow is used to measure/contol the amount of gas, then conditions need to be known and flow units standardised. Common units used:
* Metco Flow Meter Reading (FMR) (specific to equipment/process)
* SCFH (Standard Cubic Feet per Hour (70F)
* N.L.P.M or SLPM (Normal/Standard Litres per Minute (0C)
If we consider say an older 7MC system, which uses flow meter tubes and reads in Metco Flow Meter Reading (FMR) units. These units are practically meaningless outside of the equipment/material spray parameters designed for it. Conversion to standard units is possible knowing FMR, pressure, temperature and tube calibration conditions (see here) Hydrogen secondary gas is far more complicated I won't explain here (see here). More recent equipment either use flow meters (using specific parameters) designed to give direct readout in SCFH or SLPM/NLPM or mass flow meters.
Quote:Looking at conversion charts 1 cfm = 28.3 lpm thus a typical gas flow of 100 cfm = 2830 lpm right? This doesn't seem right since the A3000 only allows 100 slpm! Am I missing something with the standard part of standard litres per minute?
Yes, 1 cfm = 28.3 lpm thus a typical gas flow of 100 cfm = 2830 lpm
Are you sure you not confusing CFM with CFH? 100 SCFH is in the right ball park for plasma gas flow. 1 SCFH = 0.472 LPM so 100 SCFH = 47.2 LPM - more like it?
but wait, 47.2 LPM from above calculation really means 47.2 LPM @ 70F around normal atmosheric pressure (standard conditions for SCFH). This is not the same as SLPM/NLPM where standard conditions are 0C not 70F . So we need to apply corrections for the different conditions. So 100 SCFH = 47.2 LPM = 43.8 SLPM or NLPM.
You can use the plasma gas flow calculator to do this easily by inserting say 100 value into one of the SCFH boxes, calculate and read across.
Final point, you need to be careful where your gas flow parameters are coming from and you need to know what equipment and gun (and set-up) they are intended for. For example you can not use 3MB/7MB/9MB gas flow parameters say for a F4 gun or for that matter when using different nozzles.
Sorry for long lecture just trying to cover everybody. Hope it helps.
Dear Mr.,
I am a research scholar from India doing my Ph.D. on Plasma Spray coatings. I have gone through your comment on Jim's question. It is really benificial to me. So I like to thank you for that first. Regarding the matter I have a querry , and that is how can I corelate the pressure(in psi) with the SLPM as well as FMR. Or specifically I need to know Nitrogen flow rate in SLPM at 80 psi when the FMR are 100,125,150,175 etc without using Hydrogen. And then if I use nitrogen with those FMR with hydrogen using 10 and 20 FMR at 80 psi of nitrogen and 50 psi of hydrogen pressure, what are the SLPM values for 10 and 20 FMR hydrogen reading?
|
|
10-15-2008, 10:00 PM,
|
|
Gordon
Administrator
|
Posts: 1,802
Threads: 70
Joined: Jun 2006
Reputation:
9
| |
RE: cfm to slpm
Hi Kazi Sabiruddin
to the Surface Engineering Forum.
(10-15-2008, 08:33 AM)Kazi Sabiruddin Wrote: Dear Mr.,
I am a research scholar from India doing my Ph.D. on Plasma Spray coatings. I have gone through your comment on Jim's question. It is really benificial to me. So I like to thank you for that first. Regarding the matter I have a querry , and that is how can I corelate the pressure(in psi) with the SLPM as well as FMR. Or specifically I need to know Nitrogen flow rate in SLPM at 80 psi when the FMR are 100,125,150,175 etc without using Hydrogen. And then if I use nitrogen with those FMR with hydrogen using 10 and 20 FMR at 80 psi of nitrogen and 50 psi of hydrogen pressure, what are the SLPM values for 10 and 20 FMR hydrogen reading?
First, I suggest reading:
https://www.gordonengland.co.uk/conversion/gasflow.htm
https://www.gordonengland.co.uk/conversion/gformula.htm
and some associated links.
For nitrogen with changes in pressure only, the formula can be simplified to
Q = F x square root (P2/P1)
Q = Standard volume gas flow rate
F = Direct reading of flow meter (FMR)
P1 = Pressure of calibration gas in absolute units (50 + 14.7 for Metco 7M standard primary gas flow tube calibrated SCFH @50 psi @ 70*F)
P2 = Pressure of gas flowing in absolute units (psi value + 14.7)
I have a calculator https://www.gordonengland.co.uk/conversion/n2flow.htm which will cover 50, 75, and 100 psi settings for 7MC and SCFH to SLPM or NLPM.
In principle the same will work for hydrogen with hydrogen flow tube calibrated for SCFH H2 @ 50 psi and 70*F. BUT in the case of the older 7MC, the hydrogen flow tubes are adjusted via the inlet and not the outlet, so this makes the above formula invalid. The flow meter reading in this case is effected by the back pressure generated by the plasma system, so changes in primary gas flow, nozzle restriction etc. make a difference, while with the primary tubes this effect can be ignored.
These calculators for hydrogen flow:
https://www.gordonengland.co.uk/conversion/h2ghflow.htm
https://www.gordonengland.co.uk/conversion/h2gpflow.htm
are based on practical measurements made with a specific series of common plasma set-ups. They may be helpful in estimating hydrogen flow. Failing that, set hydrogen flow to achieve desired plasma power.
Hope that helps, but I may be assuming wrongly that your using older Metco equipment, in which case you will have to read between the lines a bit
|
|
03-19-2009, 03:11 AM,
|
|
Bostick123
Member
|
Posts: 5
Threads: 1
Joined: Jan 2009
Reputation:
0
| |
RE: cfm to slpm
Gordon,
Where wanting to do our own calibration for the Metco 9mc console, when cal-ing the Argon, HE and H2 what arethe PSI used to calibrate the flow tube. i was going by the marking on the tube 75 PSI at 70 deg for the argon, 50 psi for the H2 and 75 psi for the HE, is this correct.
|
|
03-19-2009, 10:05 PM,
|
|
Gordon
Administrator
|
Posts: 1,802
Threads: 70
Joined: Jun 2006
Reputation:
9
| |
RE: cfm to slpm
Hi Bostick123
I would base your calibration checks on the information on the tubes. I'm not sure of the 9MC ones, I know they are meant to give direct readout in SCFH for argon @ 75psi for primary. The old 7MC used primary tubes calibrated for SCFH with nitrogen @ 50psi will also give pretty good SCFH readings for argon flow when set at 75psi, but most parameters were set at 100psi so not directly reading SCFH. Most 9MC parameters call for 75 psi so making the tube reading directly in SCFH.
I would be interested to know if the primary tube is in fact calibrated with argon @ 75psi, 70F or if as I would have first suspected calibrated with nitrogen @50 psi, 70F. Should not make any real difference as they are basically the same. Helium tubes use to be calibrated with nitrogen also for 7MC.
|
|
03-19-2009, 11:50 PM,
(This post was last modified: 03-19-2009, 11:51 PM by Bostick123.)
|
|
Bostick123
Member
|
Posts: 5
Threads: 1
Joined: Jan 2009
Reputation:
0
| |
RE: cfm to slpm
(03-19-2009, 10:05 PM)Gordon Wrote: Hi Bostick123
I would base your calibration checks on the information on the tubes. I'm not sure of the 9MC ones, I know they are meant to give direct readout in SCFH for argon @ 75psi for primary. The old 7MC used primary tubes calibrated for SCFH with nitrogen @ 50psi will also give pretty good SCFH readings for argon flow when set at 75psi, but most parameters were set at 100psi so not directly reading SCFH. Most 9MC parameters call for 75 psi so making the tube reading directly in SCFH.
I would be interested to know if the primary tube is in fact calibrated with argon @ 75psi, 70F or if as I would have first suspected calibrated with nitrogen @50 psi, 70F. Should not make any real difference as they are basically the same. Helium tubes use to be calibrated with nitrogen also for 7MC.
Thanks for the Info, when i used AR to cal with at 75 PSi i got one for one, and everything seemed to come out good.
|
|
|